MRI safety and contraindications
MRI uses no ionising radiation, but the scanner room is one of the most hazardous places in a hospital. Three fields cause harm: the static magnetic field (1.5 or 3 tesla, always on) pulls ferromagnetic objects into the bore as projectiles and twists or moves implants; radiofrequency (RF) pulses deposit energy as heat and can burn skin, especially where conductive loops or wires lie; and rapidly switched gradients cause loud noise and peripheral nerve stimulation. Liquid helium adds the risk of a quench. Safety therefore depends on zoning (ACR zones I–IV), screening every person before they enter, and classifying every device as MR Safe, MR Conditional or MR Unsafe. Classic contraindications are ferromagnetic intraocular foreign bodies, older ferromagnetic aneurysm clips and some active implants, while many pacemakers are now MR Conditional and even legacy devices can be scanned under a strict protocol. In an emergency the patient is taken out of zone IV; the magnet stays on.
Patient case, mechanism, investigations, management, flashcards and exam-style questions.Open the lesson →

Key facts
- The static field is always on, even when no scan is running and even during a power cut: a quench is the only way to switch it off quickly
- Projectile (missile) injuries come from ferromagnetic objects such as oxygen cylinders, scissors, stethoscopes and trolleys taken into zone IV
- RF heating is measured as specific absorption rate (SAR, W/kg); whole-body SAR is limited to 2 W/kg in normal operating mode and 4 W/kg in first-level mode
- ACR zones: I public, II reception and screening, III restricted (control room), IV the scanner room; only screened people enter III and IV
- Device labels: MR Safe (green square), MR Conditional (yellow triangle, safe only under stated conditions), MR Unsafe (red circle with a diagonal bar)
- A history of a metal fragment entering the eye needs orbital radiographs (or CT) before MRI; a retained ferromagnetic fragment can move and blind the patient
- MR Conditional pacemakers and defibrillators need the whole system (generator and leads) to be conditional, programming to MR mode, monitoring and reprogramming afterwards
- Gradient switching produces sound above 100 dB: hearing protection is mandatory for every patient
Differential diagnosis
| Condition | Clues | Key test |
|---|---|---|
| RF burn | Localised heat or pain at a skin-to-skin loop, ECG lead, cable or metallic patch | Stop the scan; inspect and document the skin |
| Peripheral nerve stimulation | Tingling or twitching during fast gradient sequences that stops when the sequence ends | Clinical; reassure, adjust sequence |
| Claustrophobic panic | Anxiety, breathlessness and pressing the call buzzer soon after entering the bore | Clinical; stop, reassure, consider wide-bore scanner or anxiolysis |
| Cardiac device malfunction | Bradycardia, palpitations or syncope in a patient with a pacemaker or ICD | ECG monitoring and device interrogation |
| Gadolinium hypersensitivity | Urticaria, wheeze or hypotension minutes after contrast injection | Clinical; treat as anaphylaxis with IM adrenaline |
| Vasovagal syncope | Pallor, bradycardia and nausea on positioning or cannulation | Lie flat, observe; ECG if not settling |
| Displaced implant or foreign body | Sudden pain or visual loss at the site of a retained metal fragment | Stop immediately; urgent clinical review and CT |
Exam pearls
Guideline
ACR Manual on MR Safety (2024) with MHRA Safety guidelines for MRI equipment in clinical use (UK)
References
- American College of Radiology, Committee on MR Safety. ACR Manual on MR Safety. American College of Radiology 2024.
- Greenberg TD, Hoff MN, Gilk TB, et al.. ACR guidance document on MR safe practices: Updates and critical information 2019. J Magn Reson Imaging 2020.
- Medicines and Healthcare products Regulatory Agency (MHRA). Safety guidelines for magnetic resonance imaging equipment in clinical use. MHRA, London 2021.
- Indik JH, Gimbel JR, Abe H, et al.. 2017 HRS expert consensus statement on magnetic resonance imaging and radiation exposure in patients with cardiovascular implantable electronic devices. Heart Rhythm 2017.
- Russo RJ, Costa HS, Silva PD, et al.. Assessing the Risks Associated with MRI in Patients with a Pacemaker or Defibrillator. N Engl J Med 2017.
- Nazarian S, Hansford R, Rahsepar AA, et al.. Safety of Magnetic Resonance Imaging in Patients with Cardiac Devices. N Engl J Med 2017.
Related topics
By Dr Khalid Seidi (B.Sc., M.D.) · MedEd World Clinical Knowledge library. Original teaching content checked against current guidelines (November 2026); confirm the latest local guidance before clinical use. Not a substitute for clinical judgement.