MRI uses no ionising radiation, but the scanner room is one of the most hazardous places in a hospital. Three fields cause harm: the static magnetic field (1.5 or 3 tesla, always on) pulls ferromagnetic objects into the bore as projectiles and twists or moves implants; radiofrequency (RF) pulses deposit energy as heat and can burn skin, especially where conductive loops or wires lie; and rapidly switched gradients cause loud noise and peripheral nerve stimulation. Liquid helium adds the risk of a quench. Safety therefore depends on zoning (ACR zones I–IV), screening every person before they enter, and classifying every device as MR Safe, MR Conditional or MR Unsafe. Classic contraindications are ferromagnetic intraocular foreign bodies, older ferromagnetic aneurysm clips and some active implants, while many pacemakers are now MR Conditional and even legacy devices can be scanned under a strict protocol. In an emergency the patient is taken out of zone IV; the magnet stays on.

Full interactive lesson
Patient case, mechanism, investigations, management, flashcards and exam-style questions.
Open the lesson →
MRI strong magnetic field warning
MRI strong magnetic field warning Warning signs on an MRI room door: strong magnetic field, no pacemakers or metal objects beyond this point. · Tim Sheerman-Chase · CC BY 2.0 · Wikimedia Commons

Key facts

Differential diagnosis

ConditionCluesKey test
RF burnLocalised heat or pain at a skin-to-skin loop, ECG lead, cable or metallic patchStop the scan; inspect and document the skin
Peripheral nerve stimulationTingling or twitching during fast gradient sequences that stops when the sequence endsClinical; reassure, adjust sequence
Claustrophobic panicAnxiety, breathlessness and pressing the call buzzer soon after entering the boreClinical; stop, reassure, consider wide-bore scanner or anxiolysis
Cardiac device malfunctionBradycardia, palpitations or syncope in a patient with a pacemaker or ICDECG monitoring and device interrogation
Gadolinium hypersensitivityUrticaria, wheeze or hypotension minutes after contrast injectionClinical; treat as anaphylaxis with IM adrenaline
Vasovagal syncopePallor, bradycardia and nausea on positioning or cannulationLie flat, observe; ECG if not settling
Displaced implant or foreign bodySudden pain or visual loss at the site of a retained metal fragmentStop immediately; urgent clinical review and CT

Exam pearls

'It is switched off' is never true for a superconducting magnet: treat zone IV as live 24 hours a day.
Non-ferromagnetic does not mean risk-free: titanium implants do not move but conductive wires can still heat.
The commonest MRI injury is a burn, not a projectile; check for skin-to-skin contact (crossed ankles, hands clasped on abdomen) before every scan.
Most joint replacements, sternal wires, coronary stents and prosthetic valves can be scanned, usually under MR Conditional terms; check the labelling, not memory.
An unconscious patient cannot answer a screening form: look for scars, check notes and old radiographs (a scout CT or chest X-ray shows most devices) and ask relatives.
Programmable ventriculoperitoneal shunt valves can be reset by the field; check the setting after scanning.

Guideline

ACR Manual on MR Safety (2024) with MHRA Safety guidelines for MRI equipment in clinical use (UK)

References

  1. American College of Radiology, Committee on MR Safety. ACR Manual on MR Safety. American College of Radiology 2024.
  2. Greenberg TD, Hoff MN, Gilk TB, et al.. ACR guidance document on MR safe practices: Updates and critical information 2019. J Magn Reson Imaging 2020.
  3. Medicines and Healthcare products Regulatory Agency (MHRA). Safety guidelines for magnetic resonance imaging equipment in clinical use. MHRA, London 2021.
  4. Indik JH, Gimbel JR, Abe H, et al.. 2017 HRS expert consensus statement on magnetic resonance imaging and radiation exposure in patients with cardiovascular implantable electronic devices. Heart Rhythm 2017.
  5. Russo RJ, Costa HS, Silva PD, et al.. Assessing the Risks Associated with MRI in Patients with a Pacemaker or Defibrillator. N Engl J Med 2017.
  6. Nazarian S, Hansford R, Rahsepar AA, et al.. Safety of Magnetic Resonance Imaging in Patients with Cardiac Devices. N Engl J Med 2017.

Related topics

Radiation doses and risksRadiation protection & lawImaging in pregnancyPaediatric imaging & doseIodinated contrast reactionsMetformin and contrastGadolinium and NSF

By Dr Khalid Seidi (B.Sc., M.D.) · MedEd World Clinical Knowledge library. Original teaching content checked against current guidelines (November 2026); confirm the latest local guidance before clinical use. Not a substitute for clinical judgement.